Subject: Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway
To: Michael Davidsaver <[email protected]>
Date: Fri, 17 Jul 2020 14:12:24 +0000
Michael,
Thank you. Yes, there are two GPIB devices connected to that E5810. The other GPIB device, a DC calibrator, and its IOC are supposed to be offline during machine operation, but they were turned on during maintenance days and left on.
There is one record scanned at "2 second" in the DC calibrator IOC. I stopped the DC calibrator IOC. Now, the DMM HP3458A system is back to its previous "normal" state: it scans at "5 second" to read a floating point number without any problem (again, I have never made this system work faster although I tried all kinds of commands/settings).
Indeed, it only takes a few milli-second to get a reply for the command “ID?”. See below.
So, the problem seems related to the DMM itself (specifically related to the data acquisition & readout) rather than the E5810 & the network.
2020/07/17 09:11:25.614 L0 addr 6 queueRequest priority 0 not lockHolder
2020/07/17 09:11:25.614 L0 6 vxiWrite numchars 4
2020/07/17 09:11:25.616 L0 6 vxiWrite
ID?
ID?\n
49 44 3f 0a
2020/07/17 09:11:25.616 L0 addr 6 queueRequest priority 0 from lockHolder
2020/07/17 09:11:25.616 L0 schedule queueRequest timeout
2020/07/17 09:11:25.616 asynManager::portThread port=L0 callback
2020/07/17 09:11:25.616 L0 vxiSetEos 1
\n
0a
2020/07/17 09:11:25.618 L0 6 vxiRead
HP3458A
HP3458A\r\n
48 50 33 34 35 38 41 0d 0a
2020/07/17 09:11:25.618 L0 vxiSetEos 0
On 7/14/20, 10:42 AM, "Michael Davidsaver" <[email protected]> wrote:
    Yong,
    As a test, have you tried sending commands through the embedded
    web server in the E5810?
    When timing reply latency, it would be interesting to
    compare with a command which should be quick, like "ID?".
    If I recall correctly, it is also possible to
    send vxi11 commands to the E5810 itself.  I believe
    it responds to "*IDN?".  This removes the DMM
    completely from the loop.  Unless this is also slow,
    I wouldn't suspect the network.
    Also, since I should have asked previously.  Is the
    DMM the only GPIB device connected to this E5810?
Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Hu, Yong via Tech-talk
RE: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Mark Rivers via Tech-talk
Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Hu, Yong via Tech-talk
Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Michael Davidsaver via Tech-talk
Prev: Re: Pure Python IOC (CAProcess issue) Michael Davidsaver via Tech-talk
Next: Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Hu, Yong via Tech-talk
Prev: Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Michael Davidsaver via Tech-talk
Next: Re: Asyn/StreamDevice for HP3458A through Agilent E5810 LAN/GPIB Gateway Ralph Lange via Tech-talk