Subject: Re: CAEN R1470ET HVPS - Asyn + Stream Device auto connect does not recover communication
From: Jeong Han Lee via Tech-talk <[email protected]>
To: "Paduan Donadio, Marcio" <[email protected]>
Date: Tue, 1 Sep 2020 16:29:39 -0700
Hi Paduan,
  Is your ioc developed by your team or CAEN? I think CAEN has their
own embedded IOC. I saw your model also in their EPICS support list.
  Best,
  Han
On Tue, Sep 1, 2020 at 3:41 PM Mark Rivers via Tech-talk
<[email protected]> wrote:
> Hi Paduan,
> I think I might understand your problem with disconnectOnReadTimeout.  You said you are using asyn R4-32 (which is 3 years old).  You are using asyn with streamDevice.
> This is from the asyn release notes for R4-35, from March 2019.
> drvAsynIPPort
> Fixed a problem with the disconnectOnReadTimeout option. Previously it would disconnect even if pasynUser->timeout was 0. This is not logical, and meant this option could not be used with StreamDevice, because StreamDevices flushes the input by reading with a timeout of 0 to support I/O Intr scanned records.
> Mark
> From: Tech-talk <[email protected]> On Behalf Of Mark Rivers via Tech-talk
> Sent: Tuesday, September 1, 2020 5:34 PM
> To: 'Paduan Donadio, Marcio' <[email protected]>
> Cc: [email protected]
> Subject: RE: CAEN R1470ET HVPS - Asyn + Stream Device auto connect does not recover communication
> Hi Paduan,
> Ø  It looks like the equipment stops answering but Asyn still shows that the connection is opened and when it writes to the port the result is Asyn success.
> If the device crashes or otherwise becomes unresponsive without properly closing the socket it will take a long time before the client (the drvAsynTCP driver) gets informed of this.  The messages it sends just queue up in the TCP stack on either the client or the server.  This just the way that TCP works.  That is why disconnectOnReadTimeout was added to asyn, and in general it is a good thing to use.  That way whenever the device reply times out asyn will close the socket and, if autoconnect is enabled, try to reopen the connection to the device.
> Ø  If we close and open the port again, communication is reestablished.
> What do you mean by that? How did you do it?
> Ø  I’ve tried disconnectOnReadTimeout. Asyn disconnected after it, but then neither auto-connect or manual connection were able to open the connection again. The system is still in this condition as we were not authorized to reboot the IOC, yet.
> I don’t understand why the disconnectOnReadTimeout disconnection would be any different from whatever you did when you said you “close and open the port again”.
> That sounds like a truly dead device.  Can you ping the CAEN device? Can you open a manual telnet connection to it on port 1470, which seems to be the port you are using?  Those are both things you can try without restarting the IOC.
> Do you believe that restarting the IOC will fix the problem, or do you think you will need to power-cycle the CAEN?
> I don’t understand this sequence:
> 2020/08/28 14:15:55.193 hvm-ltuh-mc01:1470 write 29
> $BD:0,CMD:MON,CH:5,PAR:VSET
> 2020/08/28 14:15:55.289 hvm-ltuh-mc01:1470 read 26
> #BD:00,CMD:OK,VAL:3000.0
> 2020/08/28 14:16:00.266 HVM:LTUH:200:ASYN: exception 3, asynExceptionTraceMask
> 2020/08/28 14:16:00.283 HV0 set Eos 0
> 2020/08/28 14:16:00.323 asynManager::portThread port=HV0 callback
> 2020/08/28 14:16:00.393 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:00.472 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:00.534 asynManager::portThread port=HV0 callback
> 2020/08/28 14:16:00.604 hvm-ltuh-mc01:1470 read.
> 2020/08/28 14:16:00.657 HV0 get Eos 0
> You said “Communication was ok, until it stopped and we set some masks to diagnose:”.  But communication does not appear to have stopped before you set those masks?  It had been less than 5 seconds since the last successful read, correct?  Was there some other reason the asynTrace was turned on then?
> asynReport after using disconnectOnReadTimeout:
> Cexp@ioc-ltuh-mc02>asynReport(4)
> HV0 multiDevice:No canBlock:Yes autoConnect:Yes
>     enabled:Yes connected:No numberConnects 146
> That shows that it has connected 146 times.  Did you connect manually 146 times? That seems unlikely, so it seems that perhaps disconnectOnReadTimeout was doing what it is supposed to do, i.e. getting a read timeout, disconnecting and reconnecting?
> Mark
> From: Tech-talk <[email protected]> On Behalf Of Paduan Donadio, Marcio via Tech-talk
> Sent: Tuesday, September 1, 2020 4:49 PM
> To: Jeong Han Lee via Tech-talk <[email protected]>
> Subject: CAEN R1470ET HVPS - Asyn + Stream Device auto connect does not recover communication
> Hello, all.
> I hope you can help me to decipher a scenario with an Ethernet device, Asyn, and Stream Device. It looks like the equipment stops answering but Asyn still shows that the connection is opened and when it writes to the port the result is Asyn success. Auto-connect does not act, I guess that is because Asyn is happy with the connection. If we close and open the port again, communication is reestablished. I’ve tried disconnectOnReadTimeout. Asyn disconnected after it, but then neither auto-connect or manual connection were able to open the connection again. The system is still in this condition as we were not authorized to reboot the IOC, yet.
> Detailed data:
> Asyn R4-32
> Stream Device 2.8.9
> 2020/08/28 14:15:55.020 hvm-ltuh-mc01:1470 write 29
> $BD:0,CMD:MON,CH:4,PAR:VSET
> 2020/08/28 14:15:55.116 hvm-ltuh-mc01:1470 read 26
> #BD:00,CMD:OK,VAL:4500.0
> 2020/08/28 14:16:00.696 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:00.775 HV0 set Eos 0
> 2020/08/28 14:16:00.816 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:00.878 hvm-ltuh-mc01:1470 write.
> 2020/08/28 14:16:00.931 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:01.010 wrote 29 to hvm-ltuh-mc01:1470, return asynSuccess.
> 2020/08/28 14:16:01.090 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:01.152 HV0 set Eos 0
> 2020/08/28 14:16:01.192 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:01.271 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:01.333 HV0 addr -1 queueRequest priority 0 from lockHolder
> 2020/08/28 14:16:01.413 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:01.474 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:01.554 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:01.615 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:01.694 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:01.756 asynManager::portThread port=HV0 callback
> 2020/08/28 14:16:01.825 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:01.905 HV0 get Eos 0
> 2020/08/28 14:16:01.945 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:02.008 HV0 set Eos 2
> 2020/08/28 14:16:02.053 HVM:LTUH:200:ASYN: exception 3, asynExceptionTraceMask
> 2020/08/28 14:16:02.136 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:02.215 hvm-ltuh-mc01:1470 read.
> 2020/08/28 14:16:02.267 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:02.330 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:02.408 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:02.470 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:02.548 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:05.582 HVM:LTUH:200:ASYN: exception 3, asynExceptionTraceMask
> 2020/08/28 14:16:05.612 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:05.679 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:05.740 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:05.819 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:05.881 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:05.960 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:06.021 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:06.100 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:06.162 HV0 set Eos 0
> 2020/08/28 14:16:06.202 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:06.282 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:06.343 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:06.422 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:06.484 asynManager::portThread port=HV0 callback
> 2020/08/28 14:16:06.554 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:06.633 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:06.695 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:06.773 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:06.835 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:06.914 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:06.975 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:07.055 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:31.102 HVM:LTUH:200:ASYN: exception 3, asynExceptionTraceMask
> 2020/08/28 14:16:31.125 HV0 set Eos 0
> 2020/08/28 14:16:31.160 asynManager::portThread port=HV0 callback
> 2020/08/28 14:16:31.229 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:31.309 HV0 schedule queueRequest timeout
> 2020/08/28 14:16:31.370 asynManager::portThread port=HV0 callback
> 2020/08/28 14:16:31.440 hvm-ltuh-mc01:1470 read.
> 2020/08/28 14:16:31.493 HV0 get Eos 0
> 2020/08/28 14:16:31.533 HV0 set Eos 0
> 2020/08/28 14:16:31.573 hvm-ltuh-mc01:1470 write.
> 2020/08/28 14:16:31.626 wrote 29 to hvm-ltuh-mc01:1470, return asynSuccess.
> 2020/08/28 14:16:31.706 HV0 addr -1 queueRequest priority 0 not lockHolder
> 2020/08/28 14:16:31.786 HV0 set Eos 0
>     enabled:Yes connected:Yes numberConnects 3
>     nDevices 0 nQueued 49 blocked:No
>     asynManagerLock:No synchronousLock:Yes
>     exceptionActive:No exceptionUsers 170 exceptionNotifys 0
>     traceMask:0xb traceIOMask:0x7 traceInfoMask:0x1
>     interposeInterfaceList
>         asynOctet pinterface 0x2e48048 drvPvt 0x2930310
>     interfaceList
>         asynCommon pinterface 0x2d5175c drvPvt 0x2926f60
>         asynOption pinterface 0x2d51768 drvPvt 0x2926f60
>         asynOctet pinterface 0x2927040 drvPvt 0x2926f60
>     Port hvm-ltuh-mc01:1470: Connected
>                     fd: 16
>     Characters written: 31886349
>        Characters read: 18628310
> Asyn report a few minutes later:
> interfaceList
>                fd: 16
>     Characters written: 31887627
> After manually reopening the connection, the device restart sending characters (this was in another day):
> 2020/08/31 12:11:57.446 hvm-ltuh-mc01:1470 write 29
> $BD:0,CMD:MON,CH:0,PAR:VSET
> $BD:0,CMD:MON,CH:0,PAR:VSET\r\n
> 24 42 44 3a 30 2c 43 4d 44 3a 4d 4f 4e 2c 43 48 3a 30 2c 50
> 41 52 3a 56 53 45 54 0d 0a
> 2020/08/31 12:11:58.600 hvm-ltuh-mc01:1470 write 29
> $BD:0,CMD:MON,CH:5,PAR:VMON
> $BD:0,CMD:MON,CH:5,PAR:VMON\r\n
> 24 42 44 3a 30 2c 43 4d 44 3a 4d 4f 4e 2c 43 48 3a 35 2c 50
> 41 52 3a 56 4d 4f 4e 0d 0a
> 2020/08/31 12:11:59.765 hvm-ltuh-mc01:1470 write 29
> $BD:0,CMD:MON,CH:1,PAR:VSET
> $BD:0,CMD:MON,CH:1,PAR:VSET\r\n
> 24 42 44 3a 30 2c 43 4d 44 3a 4d 4f 4e 2c 43 48 3a 31 2c 50
> 2020/08/31 12:11:59.926 hvm-ltuh-mc01:1470 read 26
> #BD:00,CMD:OK,VAL:0800.0
> #BD:00,CMD:OK,VAL:0800.0\r\n
> 23 42 44 3a 30 30 2c 43 4d 44 3a 4f 4b 2c 56 41 4c 3a 30 38
> 30 30 2e 30 0d 0a
> 2020/08/31 12:12:00.122 hvm-ltuh-mc01:1470 write 29
> $BD:0,CMD:MON,CH:0,PAR:STAT
> $BD:0,CMD:MON,CH:0,PAR:STAT\r\n
> 41 52 3a 53 54 41 54 0d 0a
> 2020/08/31 12:12:00.345 hvm-ltuh-mc01:1470 read 25
> #BD:00,CMD:OK,VAL:00001
> #BD:00,CMD:OK,VAL:00001\r\n
> This is how Asyn report looks like one things are going well:
>     enabled:Yes connected:Yes numberConnects 4
>     nDevices 0 nQueued 3 blocked:No
>     traceMask:0x1 traceIOMask:0x1 traceInfoMask:0x1
>     interfaceList
>                     fd: 16
>     Characters written: 32245553
>        Characters read: 18922078
> Dbpr when things are bad:
> Cexp@ioc-ltuh-mc02>dbpr("HVM:LTUH:200:BoardName","4")
> ACKS: INVALID       ACKT: YES           APST: On Change     ASG :
> ASP : PTR 0x3651f20 BKLNK: ELL 1 [0x2ea1e00 .. 0x2ea1e00]   BKPT: 00
> DESC: BoardName     DISA: 0             DISP: 0             DISS: NO_ALARM
> DISV: 1             DPVT: PTR 0x35bbea0 DSET: PTR 0x2e4882c DTYP: stream
> EVNT:               FLNK: CONSTANT
> INP : INST_IO @caenN1470.proto readstring(0,BDNAME,HVM:LTUH:200) HV0
> LCNT: 0             LSET: PTR 0x2fa7680 MLIS: ELL 2 [0x39950b8 .. 0x3994f00]
> MLOK: 03 50 22 90   MPST: On Change     NAME: HVM:LTUH:200:BoardName
> NSEV: NO_ALARM      NSTA: NO_ALARM      OLDSIMM: NO         OVAL: R1470ET
> PACT: 1             PHAS: 0             PINI: NO            PPN : PTR 0x0
> PPNR: PTR 0x0       PRIO: LOW           PROC: 0             PUTF: 0
> RDES: PTR 0x2be1910 RPRO: 0             RSET: PTR 0x2e498e0 SCAN: Passive
> SDIS: CONSTANT      SDLY: -1            SEVR: INVALID       SIML: CONSTANT
> SIMM: NO            SIMPVT: PTR 0x0     SIMS: NO_ALARM      SIOL: CONSTANT
> SPVT: PTR 0x0       SSCN: <nil>         STAT: TIMEOUT       SVAL:
> TIME: 2020-08-31 13:09:47.237075064     TPRO: 0             TSE : 0
> TSEL: CONSTANT      UDF : 0             UDFS: INVALID       VAL : R1470ET
> Dbpr after using disconnectOnReadTimeout:
> PACT: 0             PHAS: 0             PINI: NO            PPN : PTR 0x0
> SPVT: PTR 0x0       SSCN: <nil>         STAT: COMM          SVAL:
> TIME: 2020-08-31 13:14:51.686086457     TPRO: 0             TSE : 0
>     nDevices 0 nQueued 28 blocked:No
>     interfaceList
>     Port hvm-ltuh-mc01:1470: Disconnected
>                     fd: -1
>     Characters written: 34072265
>        Characters read: 20373454
> Turning off disconnectOnReadTimeout and trying to manually disconnect didn’t work anymore:
> 2020/08/31 14:32:29.651 Can't connect to hvm-ltuh-mc01:1470: Connection timed out HV0 -1 autoConnect could not connect
> 2020/08/31 14:33:44.717 Can't connect to hvm-ltuh-mc01:1470: Connection timed out HV0 -1 autoConnect could not connect
> 2020/08/31 14:44:13.743 HVM:LTUH:200:ASYN: queueRequest failed
> 2020/08/31 14:44:13.745 HVM:LTUH:200:ASYN: queueRequest failed
> 2020/08/31 14:44:15.565 HVM:LTUH:200:ASYN: queueRequest failed
> 2020/08/31 14:44:15.567 HVM:LTUH:200:ASYN: queueRequest failed
> 2020/08/31 14:44:21.023 HVM:LTUH:200:ASYN: special queueRequest timeout
> 2020/08/31 14:44:27.077 HVM:LTUH:200:ASYN: special queueRequest timeout
> 2020/08/31 14:45:19.767 Can't connect to hvm-ltuh-mc01:1470: Connection timed out HV0 -1 autoConnect could not connect
> 2020/08/31 14:46:34.831 Can't connect to hvm-ltuh-mc01:1470: Connection timed out HV0 -1 autoConnect could not connect
CAEN R1470ET HVPS - Asyn + Stream Device auto connect does not recover communication Paduan Donadio, Marcio via Tech-talk
RE: CAEN R1470ET HVPS - Asyn + Stream Device auto connect does not recover communication Mark Rivers via Tech-talk
Prev: RE: CAEN R1470ET HVPS - Asyn + Stream Device auto connect does not recover communication Mark Rivers via Tech-talk
Next: Re: running gateway on mac Ralph Lange via Tech-talk
Next: change MACRO with text-entry value in CSS Phoebus Peter Leban via Tech-talk